Hi guys- sorry for the much longer than expected delay on getting back with the results.
These cubes do work from 400-700nm (and by the looks of it beyond that if you have a use), and the extinction ratio is quite good (>99% for 600-400nm, >95% for 600-700nm).
The 'insertion loss' (for lack of a better word) leaves something to be desired, and is as much as 10% in some of the shorter wavelengths, which I believe is due to absorption in the glass. Do we know what these are made out of?
The results are summarized below, divided into 2 groups of testing.
Group 1 - High Power Tests
Code:
wavelength polarization deflected passed input power %pass %deflect
405nm v 0.2 28.3 30.8 91.88 0.65
605nm h 28 0.6 30.9 1.94 90.61
445nm v 0.05 47.6 50.7 93.89 0.09
445nm h 48.2 0.01 50.9 0.01 94.7
660nm v 0.15 9.25 9.53 97.06 1.57
660nm h 0.2 9.1 9.45 96.3 2.12
This test was performed by shining a diode laser into the PBS, with both vertical and horizontal polarizations. The output powers (in mw) were measured using a photodiode based power meter (coherent fieldmaster w/ lm2 head), and the calculated transmission/reflection are plotted.
Group 2 - Low Power Tests
Code:
wavelength polarization deflect pass original %pass %defect
676nm v 4 1800 1800 100 0.22
676nm h 1800 91 1850 4.92 97.3
647nm v 71 7300 7300 100 0.97
647nm h 7600 67 7600 0.88 100
520nm v 16 1400 1500 93.33 1.07
520nm h 1300 0.2 1350 0.01 96.3
524mn v 5 390 400 97.5 1.25
524nm h 340 0.1 340 0.03 100
488nm v 50 2900 3000 96.67 1.67
488nm h 3200 3 3300 0.09 96.97
476nm v 2 310 320 96.88 0.63
476nm h 310 0.4 320 0.13 96.88
This test was performed using a whitelight argon + pcoam, with physically rotating the PBS to compare V/H polarization. The powers are in arbitrary units (I did not bother dialing the wavelength in for each one, but its roughly 1uw per count). Also note- due to the low power available, I could not take accurate readings on the transmitted direction, so most of the readings are 100% or close to it. Refer to the high power test section for accurate readings regarding the losses in he transmitted direction.