Abstract consoles existed LONG before KPPS... It dawned on me last night how to explain this...
Forget KPPS!
!
!
!
Ok, When driven by graphics software, a Galvo operates in a ballistic mode. It is forced to its maximum performance by a carefully designed waveform designed to push the galvo to its limits. That performance is measured in industry using a standard waveform, the repetition rate of which is measured in KPPS.. In audio terms, the galvo-amplifier combination operates as its own reconstructive filter for the digital waveforms, which is why a 30 KPPS waveform really only has actual deflection components to 2.4 KHz or so, plus harmonics at the edges.. The test pattern pushes the Galvo to its performance limits by design. All graphics software does this, but classical abstract machines DO NOT. They have NO CPU processor to do it!!
~
~
KPPS divided by number of points in the image is the FUNDAMENTAL rate that the waveform repeats, in Graphics...
~
~
Forget all that, and read this:
~
When driven by a abstract generator outputting sine or triangle waves at up to 85% of its first resonant frequency, a Galvo runs in LINEAR mode and just tries to "keep up" with the waveform. There are no KPPS in that mode... There is no need to add corner points, anchor points, stretching points, or to linearize the velocity... Its just following the waveform like an oscilloscope trace would.
~
Keep the Galvo system filtered below 85% of its first mechanical resonance and your fine...
~
First resonance measured last night on a poorly tuned "30K" 6800 scanner with a 6850 amplifier with no performance enhancing notch filters installed on the amplifier, and standard mirror set was ~1600 Hz... 85% of that was 1350 Hz, or .707 times that is 1130 Hz... So have a switch selectable low pass at 500 Hz and 1000 Hz and you'll hit your target... Make sure it has low phase shift in SPICE simulation. Filter both axis exactly the same...
~
Shaft resonances that the scanner amplifier can't damp are what will overheat the Galvo fast, or cause its amplifier to oscillate...., and the edge of square wave from a music synth will contain all those frequencies... Driven hard at 11 KHz for example, the amp entered "protective shutdown" in a fraction of a second... Most modern scanner amps have notch filters in the feedback loop at the first and second mechanical resonances... So they can do the "digital" graphics..
~
~
Note: Measuring the mechanical resonance requires detuning the scanner amp feedback loop, as documented in the factory manual. It is not something an average user can do..
~
I can show you in published literature where the scanner has ballistic and linear modes, it will take a few days to dig up a paper by Brosens... One of the early galvo designers.... One of the goals of a good scanner amp is to not exhibit a noticeable change between either mode...
~
No need for a variable filter, that will just allow users to pass resonances. Sad thing about scanner resonances...At low amplitudes they add a cool looking "effect" to the abstract and users tend to dwell there, because the scanner uses less energy at first...
~
~
Steve