Brad if you have schematics that would save me a lot of time right now...
More from my friend in Germany...
The ignition box from the ALC 909Z generates everything it needs from the 300V (tube) open circuit voltage of the power supply unit. Only the control input, which is internally separated by an optocoupler, has to be connected to an external 9-15V. The input is labeled "L + / L-" on the circuit board, on my box there was a coaxial cable with a BNC connector at the end.
If you add a 9V block battery and a button, the whole box can be operated quite well as a stand-alone device.
Debounce the signal to the optocoupler. make it a sharp square wave. Weak or oscillating control signals have a habit of causing ignition down the gas return bores in the tube. This is by generating too weak a pulse.
Steve