    Default Intracavity beam skewing due to thermal lensing?

    I've been working on a large DPSS green for a couple of years and am trying to implement a design that uses passive Q-switching to increase the intracavity intensity for green conversion. The design I'm trying is a z-fold configuration where two concave mirrors form a focal point to provide the right intensity to the saturable absorber, a fairly large pump mode volume, and about a 2x reduction in area going into the KTP. I've been using the free software Rezonator to work out the cavity and the software predicts its stable down to about 30cm of thermal lensing in the vanadate crystal. Here is a diagram of the resonator design:

    Click image for larger version. 

Name:	2023 Design Schematic.png 
Views:	2 
Size:	59.2 KB 
ID:	60239

    Attachment 60238
    I have this lasing on my workbench in CW mode -- all the mirrors are in place but there is no Cr:YAG or KTP crystal in the cavity. The problem I'm having is there is a very narrow range of pump powers where it lases. If I turn up the pump power past this range lasing stops. I think this is thermal: If I start with the pump up I'll get lasing and then the beam will fade, flicker and finally go out within a few seconds. I don't have a TEC on the vanadate but I do have it in a copper block and the block doesn't get warm so I don't think adding a TEC will help. A thermal camera on the vanadate shows it's only about 80ºF. I think something about this cavity is magnifying the thermal lensing of the vanadate.

    I've done some simple thermal lensing calculations and they show the vanadate should be nowhere close to the 30cm of lensing needed to lose stability. This is a pretty long cavity design (about 650mm) and there is some astigmatism caused by the beam hitting the concave mirrors at an angle (max about 10º). I don't know how much of a difference that makes (Rezonator shows the astigmatism actually decreases with more lensing).

    I had a theory that maybe the lensing is causing the beam angle to change in the cavity and this is causing it to walk off the pump mode overlap. As a test I turned the pump up until the laser stopped, and then adjusted the X-axis of the pump and viola! I got the beam back. There is increased deflection as I increase the pump, and if I back the pump power off the deflection decreases so there is really only a narrow sweet spot of pump power where it works, but I can tune that spot for any power.

    As lensing increases the focal point between the concave mirrors shifts to the left. I assume my alignment is not perfect enough in the vanadate and the beam is not completely parallel. But when I aligned this I used several 1mm pinholes and the beam seems to be going straight through the vanadate. I don't know how I could make this more precise.

    Can anyone think of any other issues here that may be causing the beam angle to walk in the vanadate as the power is increased? Any idea how to eliminate this?

    Thx,
    Brian
    Last edited by brianpe; 05-13-2023 at 14:32.

  2. #2
    Join Date
    Oct 2010
    Posts
    41

    Default

    Take your Thermal cam and check the folding/focusing mirrors. If you see a hot one -> there is your problem.

  3. #3
    Join Date
    Jun 2017
    Location
    Seattle, WA
    Posts
    47

    Default

    Quote Originally Posted by Farbe View Post
    Take your Thermal cam and check the folding/focusing mirrors. If you see a hot one -> there is your problem.
    Great idea. I tried this, but all the mirrors show as cold. I was also wrong in my assessment that this can be fixed by walking the pump beam. Yes, it can, but after re-aligning again I still have the problem but walking the beam no longer works. So it is still a lensing issue.

    One thing I tried is to shorten the lower arm where the KTP goes. I shortened from 98mm to 73mm. This increases tolerance from lensing and allows lensing down to 25cm (from 32). When I do this I can increase the power more. It still fails far before where I predict but I think this shows some predictions hold.

    So three things:

    First I designed this cavity to have a largish gain mode area to reduce pump lensing. It's 430µm, but elliptical due to the concave mirrors -- the larger radius is 550µm. I've read that while increasing the area does reduce thermal lensing, it makes the cavity more susceptible to lensing. Is this just too big? The mode diameters don't change too drastically as the lensing increases.

    Second is that maybe I'm estimating lensing wrong? Vanadate has different dn/dK values for different polarizations. My pump is a multimode fiber that causes random polarization. I'm estimating this by calculating lensing for 50% e polarization and 50% o polarization and taking the average. I've also tried adding these together and this is still double what I'm seeing on the bench.

    Finally, does cavity length make a big difference here? I've built several cavities and never had such an extreme lensing problem, but they have all been shorter and have a smaller gain mode.

    I'm only driving this with about 5 watts of pump before it stops....I find it surprising that this would yield 32cm of lensing so early on.

  4. #4
    Join Date
    Oct 2010
    Posts
    41

    Default

    If I am not mistaken, at 5W heat load and 200u beam radius you should expect a thermal lens with focal lengths of around 85mm or 30mm depending on thermal optical coefficient used.
    The low thermal conductivity is worsening that by quite a bit.
    Its only 5W/m/k.

  5. #5
    Join Date
    May 2009
    Location
    Sacramento, California
    Posts
    608

    Default

    I have had lensing issues with my 5 watt green YAG in the past. it was an external aluminum mirror that had a protective coating on it that would cause the beam to bloom or diverge at 4 to 5 watts right off the head. I swapped to an uncoated mirror and the problem went away.
    Will there be three phase!!!!

  6. #6
    Join Date
    Nov 2008
    Location
    Cleveland Ohio
    Posts
    2,599

    Default

    you might try skewing the crystal in the cavity. The focus of the gauss point in the beam might walk a tiny bit and that might fix it just by moving the point of focus in the crystal. It likely will drop output power a little too. I'm thinking this might select a slightly different phase and or mode. You could also try temp tuning and or rotating the crystal to see if there is a better phase match. There are usually four places where (n1+n2)/2 is met. KERR might be self focusing...is self focusing and could skew the acceptance angle where efficient phase matching occurs. You might put a rotator in the cavity and see if changing the phase of the pump beam might help a better phase match while reducing self focusing. You have to satisfy the collinearity of the photons or no SHG. Depending on how the birefringence is located in the crystal and the cut you might have to play a bit. I'd expect a bought crystal would be labeled as such.

    This paper gets the points pretty well. https://authors.library.caltech.edu/2871/1/ANDao84a.pdf
    Look at figure three. You can do the math but its all there in he text.

    I missed the passive qSwitch element. Does the beam walk when the qswitch dumps? iS it stable in CW?
    Last edited by kecked; 05-23-2023 at 12:05.

  7. #7
    Join Date
    Jun 2017
    Location
    Seattle, WA
    Posts
    47

    Default

    Thanks all for the replies. Vidal -- I know exactly what you're referring to with the silver front surface mirrors. I don't use these inside the cavity but I use them outside the cavity to steer the alignment laser and yes, when I reflect an IR beam from them it does some pretty unruly things when I turn the power up. I don't see this with the higher quality mirrors I'm using intracavity.

    Kecked -- Thx for the paper. I have been reasonably careful to ensure alignment between the cavity mode and pump, but it was very useful reading this paper to understand more how things can go awry if the two are off. Something to be careful about -- I've tried to keep them aligned but as the cavity alignment progresses with more mirrors I am probably less stringent than I should be here.

    As I get the cavity going I've removed both KTP and Cr:YAG so this is running only in CW. I've read that while larger mode diameters do reduce thermal lensing they cause the cavity to be more sensitive to thermal lensing. I haven't found a rigorous analysis of this -- more anecdotal. But anyway I made some small changes to the cavity design that reduces the mode area in the vanadate by a small amount so it's around 400µm. I re-aligned to these new dimensions and now I can bring the cavity up to the power limit of my bench setup and I never lose stability. I don't know if it was the mode area change that did it, or just better alignment.

    Next step is to insert the KTP, then the Cr:YAG. To do this I need to swap out one mirror: one of the mirrors is currently HR1064, HR 532 and needs to be swapped to HR1064 HT 532. I didn't align with this mirror in place initially because I need a pretty bright light source to get this many mirrors aligned. I'm using a small 532 DPSS to do the alignment (I also have a 5mw HeNe but it is just too dim). The HT 532 is also AR coated for 532 and this leaves too few photons to see. So far I haven't been successful swapping this mirror out. The swap just isn't aligned well enough. I'm considering buying a low power 1064nm laser for alignment, but not thrilled about aligning the cavity staring at a monitor.