Oxygen adsorption is strong in some parts of the tuning range. You can actually see the dip in lasing power as you pass through an O2 or H2O line. Can also be sensitive to humidity. Leave the ti-saph rod in the mount. Its factory aligned. On "my" Mira long ago, the post-doc had rotated the rod about it's axis, it took me a great deal of time to find the optimum spot over two weeks.
Do not swap optics, make sure they are numbered as you remove them, as they can be tailored to each laser, and often the parts numbers are not in the manual.
The Birefringent tuner has multiple stacked ranges, there is one optimum one, and I'd find it and leave it there.
Not a great idea to change BRF angle relative to the beam.
As Kecked said, once you find the pump sweet spot, and the optimal temperature for the chiller, Ti-Saph wants to lase and lase CW.
Silicon CCD cameras with a 10cm to 2 meter target distance C mount lens, of the kind used for filming cash registers for security, plumbed into a large monitor makes working on MIRA a dream. I did any course alignment with a 785 nm Laser diode. The main ring is easy to align.
Auto-start with the clear bar on the moving actuator open loop galvo, is touchy to adjust, Some times , as Kecked said, its just easy to manually tap so it chops the beam with it's self resonance.
Somewhat dust sensitive, so good to tune it with the cover on.
I have no experience with the built in REGEN.
MIRA IS A FUN LASER !!!
Steve