OK, I need to look at an MRA field install sheet before I comment, but your tube I-V plot actually is not that bad. That will take a day or two.
How often do you run it? How many hours has it had in the last year?
Correct sequence is to energize the ready solenoid, fill the intra solenoid chamber, turn off the ready solenoid, disconnect it from any power source, then energize the fill solenoid. We then close and disconnect the fill solenoid. Both solenoids must never be open at the same time because the resivoir is at nearly atmospheric pressure. In reality most of us use low voltage AC and "buzz" the fill solenoid to slow the flow.
Don't fill it yet. I'm not sure you you need a fill. In fact you may need to use the laser some. The data is borderline for a fill, but may indicate and overpressure from the natural leakage of the reservoir.
Optics peaked and clean? Do you have access to a laser power meter?
Your tube is Pure Argon, Pure Krypton, or White-light?
Where are you located?