One quick question...
It seems like at idle with the field turned all the way down I only get the 488 line and just a tiny bit of the 514nm line. Is this normal? As soon as the field or current goes up I get all 6 lines again. I was just a bit curious.
Also, is the magnet AC powered or DC powered? I ask because the 514 line almost seems to be 'pulsing' until the field or current is turned up. I cant see the pulsing until I used a diffraction grating to 'move' the beam around. The 488 line seems fine. I would assume that this could be because the magnet is ac powered and running at 60hz it would cause the line to pulse until it has more power.
the magnet is DC. You are just barely at threshold and 514 is trying to lase. 488 is a higher gain line. Ya might be a bit high pressure but until we get some data... ya just dont know. So you can only run it at minimum current?
Remember I said get her running 35 amps field max 237 -243 vdc across tube. I thought I sent you the testing parameters...may be not.
Nice little argon ya got there!
Pat B