Hi Dan,
I doubt the issue is tuning since it appears to be occuring on both galvo's at the same time. It's been a while since I've messed with a set of DT-25's but I remember that they had a single amp that handled X and Y but it was essentially 2 amps on 1 card. If I remember correctly, the power supply is +/-24V and the amp has an onboard 7815 and 7915 regulators. So regardless of the ripple you're seeing on the main supply, the amp drops this down to +/-15V (IIRC). You could look at the output of the regulators to make sure they are stable. I see from the pic above that you have a scope, does it have an X/Y setting? Have you looked at the output of the FB3? What all is between the FB3 and the scanner amp? I'm guessing at least an ILDA cable to a DB25 in the projector and wires directly to the amp? Do you have ILDA pin 25 connected to the amp or to the power supply?