I'm having an interesting ION issue and I can't come up with a good reason for it. It's a Reliant 300WC with around 160 hours. The head was last serviced in 2007, originally manufactured in 2003. When it first starts it has a great color balance though it does lean a bit red. If I let it run for 6-8 hours at around 9-10 amps it will shift largely to red, leaving the other lines very weak. If I turn it off for 6-8 hours and then turn it back on, it again starts with a relatively nice balance, then shift red within about 30 minutes. If I leave it off for a couple weeks I can get a good 6-8 hours at 9-10 amps before it shifts red again.
When I say it shifts red, yellow is almost completely gone, the 2 green lines are both very weak and with a semi strong blue line. To me, from reading thru the laser FAQ, this would indicate high krypton pressure, but why start with such a balance of color only to shift to high krypton pressure with use. I'm not understanding why the Vdrop is going up unless maybe there's a contaminant in there or perhaps I just need to run the hell out of it to get the krypton down.
If it helps, at 9amps tube voltage is around 180VDC when the colors are balanced. When it's shifted red the tube voltage is around 193VDC. When this occurs I have ramped the current up to max, 12A, and the voltage drop went up to 205VDC.